Home Forum Support NaN on te map (si8b tube)
  • #5305
    OrtoSystem
    Yes, i see the web page.
    If i download the data from dashboard the value are correct.
    OrtoSystem
    #5308
    Hi guys,
    Indeed the new tube needs to be defined on the server side too, I am perfectly ok to allocating 0xC for this excellent pancake tube.
    I do have one and can run the tests to determine the approximative conversion from CPM to eq dose rate, but only in a few weeks from now.
    So instead, we might put some approximative values in place, if @ortoSystem can help me with that. I would need background radiation readings on two units, one with the standard SBM20 tube and next to it the unit with SI8B.
    #5316
    OrtoSystem
    Hi Radhoo,
    I did these two tests:
    All cpm are a mean on an hour.
    I run 1 hour of measurements with the SI8B and SBM20 and I found a conversion factor of 0.0011081.
    SBM20 = 16.27 cpm = 0.1028 usv/h (conv. 0.006315)
    SI8B = 92.77 cpm = conversion factor 0.0011081 (I assume the same value measured with SBM20 0.1028 usv/h)
    I run a parallel measurement with a bgeigie nano kit and I found a difference on the mean values on one hour of 0.00546 usv/h.
    SI8B = 92.75cpm (conv 0.0011081)= 0.1027825 usv/h
    Bgeigie= 32.5cpm (conv 0.002994) = 0.097324 usv/h
    So I think that a conversion factor of 0.0011081 can be a good value, what do you think?
    OrtoSystem
    #5339
    Thank you for this excellent contribution.
    SI8B was added with a linear factor of 0.001108
    It will be added to the KIT1 repo soon too!
    Direct unit link: https://www.uradmonitor.com/?open=13000208&sensor=cpm
    Merry Christmas!